HiPALS needle position z-axis

Hello everyone,

we are running a 1260 Infinity with a HiP ALS. The needle needs to be adjusted for the z-axis / needle-down position, since about 40 µl of sample remain in the insert of the vial.

We are using the LabAdvisor software.

Does anybody can provide a way how to calibrate the needle-down position?

Parents
  • Hey, in Lab Advisor you cannot adjust this. You can do it in the method in the chromatographic data system (CDS) as it is part of the LC driver. Go to the G1367E method, then to the Advanced tab, there you can adjust the "draw position". See the screenshot. A negative value will move the needle down lower. Be careful not to crash the needle into the inserts. 

Reply
  • Hey, in Lab Advisor you cannot adjust this. You can do it in the method in the chromatographic data system (CDS) as it is part of the LC driver. Go to the G1367E method, then to the Advanced tab, there you can adjust the "draw position". See the screenshot. A negative value will move the needle down lower. Be careful not to crash the needle into the inserts. 