Hello,
Have you faced some RT differences between Agilent and Waters? I´m facing 7 min of differece, it is normal or could be another kind of problem?
Agilent RT = 33,3 min
Waters RT = 26,6 min
Hello,
Have you faced some RT differences between Agilent and Waters? I´m facing 7 min of differece, it is normal or could be another kind of problem?
Agilent RT = 33,3 min
Waters RT = 26,6 min
What size column, and what flow rate on each system? I'm assuming the same column is used on both systems.
Hey Julia,
It sounds like it is a difference in system volumes and the delay volume it takes for the gradient to reach the column. Maybe you have different types of pumps (binary versus quaternary) or maybe different samplers (flow-through needle versus fixed loop). These hardware differences create volume differences which result in different retention times.
Are you running a gradient? If isocratic, the elution time should be based mostly on the flow rate, assuming the same solvent composition. If a gradient, the comments from kbusch will apply.
Are you running a gradient? If isocratic, the elution time should be based mostly on the flow rate, assuming the same solvent composition. If a gradient, the comments from kbusch will apply.
It is a gradient!! Manu thanks for reply