Hello,
Have you faced some RT differences between Agilent and Waters? I´m facing 7 min of differece, it is normal or could be another kind of problem?
Agilent RT = 33,3 min
Waters RT = 26,6 min
Hey Julia,
It sounds like it is a difference in system volumes and the delay volume it takes for the gradient to reach the column. Maybe you have different types of pumps (binary versus quaternary) or maybe different samplers (flow-through needle versus fixed loop). These hardware differences create volume differences which result in different retention times.