Agilent 1200 G1312B SL Pressure Issue

Hi everyone,

This is our lab first time using a G1312B SL Bin Pump. The pressure we are getting is quite unusual.

We have replaced the seals, the frit, outlet ball valve and tested other pieces. Now, after replacing the damper, the pressure is stable yet still inconsistent with what we are used to.

For instance, open valve, 5ml IPA, 100% A1, we are getting >40 bar. With close valve, 2ml IPA A1, we are getting 33-34bar stable.

Any thoughts would be appreciated.

Thank you,

Parents Reply Children
  • Thanks for confirming. At this point the cause may be within both pump heads and/or between these and the purge valve (inclusive).
    Since you have had this issue for a while and have already replaced a few parts, however the issue still persists (with the possibility of more than one part being the cause), it is best to contact your local Agilent Customer Support Team. You can share the pump model number and serial number + description of the issue and parts you have already replaced. They will be able to help you diagnose this more closely.