Wavelength Calibration test aborted by system - HPLC 1100 DAD detector

I have just replaced both the tungsten and deuterium lamps in the DAD of my 1100 HPLC and when I attempt to run the wavelength calibration test in lab advisor, it won't complete the test and the status is aborted by system. The lamps have been on for the required time and the flow cell is removed during the test.

There is no further information about why the test is aborting - does anyone have any further insight

Parents Reply Children
  • No, no error codes in the results file or the log book, the alpha and beta deviation are very high though

  • There could be hardware issue like motherboard but before that do lamp intensity test then try wave length calibration again.

  • The lamp intensity test fails - does not get aborted but fully fails. Any information on how I can tell if it's a motherboard issue? Thanks 