Hi Carol,
When you say RT is shifting, is it longer and constant, shorter and constant or moving back and forth?
What is your pump model and are you running a gradient?
Checking the flow rate as previously mentioned is a good idea but if your pressure is steady under isocratic conditions, then it a good assumption your check valves are okay. At that point, you can still have a leak so checking the flow will confirm whether to check for a leak or not.