HPLC ISSUE

Hi,

I have an issue 1100 HPLC. The retention time is shifting. I changed the station pump, the needle, the needle seat, the column, the seal rotor, and the guard. The issue continues. Some ideas and recommendations.  

Thanks

Parents
  • Hi,

    1. Measure flow before column and then from drain line by using digital liquid flow meter or any measuring glass vial with stop watch.Then see flow is okay or not.

    2.

    Check out valve of pump may malfunctioning so try to change it.

    3.

    Variation of temperature of column compartment will vary RT.

    4.

    Replaced column is same as previous or not.

    If all above is okay then do calibration again to see the result.

    Thanks

Reply
  • Hi,

    1. Measure flow before column and then from drain line by using digital liquid flow meter or any measuring glass vial with stop watch.Then see flow is okay or not.

    2.

    Check out valve of pump may malfunctioning so try to change it.

    3.

    Variation of temperature of column compartment will vary RT.

    4.

    Replaced column is same as previous or not.

    If all above is okay then do calibration again to see the result.

    Thanks