system pressure test on a G1312B

On a G1312B would there be a benefit to running the HPLC lab Advisor system pressure test on both channels A and B?

Why are two channels given as an option?

Is there an option to run the pressure test on different channels for a quaternary system? 

Thanks

Parents
  • 1) Yes, you are sure there is not leak in any pump head

    2) For performing the test to channel A, B or both, if you perform the test in 1 channel it doesn't test both head pumps

    3) No, quaternary system has just 1 pump head; even it has 4 solvent tubings (channels) all they are before pump head in the flow path, and they are at atmospheric pressure, the mixture of solvents occurs also at atmospheric pressure, the solvent mixture enters to pump head and then it is pressurized, so in a quaternary system it is not important which solvent channel you select for running the test.

Reply
  • 1) Yes, you are sure there is not leak in any pump head

    2) For performing the test to channel A, B or both, if you perform the test in 1 channel it doesn't test both head pumps

    3) No, quaternary system has just 1 pump head; even it has 4 solvent tubings (channels) all they are before pump head in the flow path, and they are at atmospheric pressure, the mixture of solvents occurs also at atmospheric pressure, the solvent mixture enters to pump head and then it is pressurized, so in a quaternary system it is not important which solvent channel you select for running the test.