System diagram not showing up

Got a weird problem today.  I have an LC1100 with chemstation that was working fine until today.  When I launched it, the chemstation loaded up but it the system diagram didn't load.  If you go to view->system diagram, it won't load that way either and on the left-bottom of the screen it will give an error 'A register is empty'.  Nothing has been changed on the computer system.  I can connect to it and control it through lab advisor.  Any advice/ideas?

Parents
  • Hello JohnB,

     

    Without LOG files or an error screenshot (e.g. autostart macro / error code number), the possibilities are numerous.

     

    here are some ideas:
    1) Reg file rename
    Exit LC Chemstation
    Open the instrument directory: "C:\Chem32\x" (x corresponds to instrument no.)
    There are 2 files that can be responsible: Config.reg & Conf_Off.reg.
    Rename them by adding the extension .old or backup.
    NOTE:
       * Be aware that for LC applications the config.reg file contains the information for the columns. On systems without column tags the column data has to be entered manually.
       * The REG files are rebuild after the next successful start of an instrument session.

    Restart Chemstation & the error could already be fixed.


    2) other steps, if (1) did not help:
    a) take a screenshot of the error
    b) which ChemStation version do you use? (menu: help> about)

     

    copy here additional files:
       c) c:\WINDOWS\ChemStation.ini
       d) instrument LOG file:
                 C:\Chem32\x\TEMP\INSTRx.LOG where X is the instrument number
          e.g. C:\Chem32\1\TEMP\INSTR1.LOG

     

     

    regards, M.

Reply
  • Hello JohnB,

     

    Without LOG files or an error screenshot (e.g. autostart macro / error code number), the possibilities are numerous.

     

    here are some ideas:
    1) Reg file rename
    Exit LC Chemstation
    Open the instrument directory: "C:\Chem32\x" (x corresponds to instrument no.)
    There are 2 files that can be responsible: Config.reg & Conf_Off.reg.
    Rename them by adding the extension .old or backup.
    NOTE:
       * Be aware that for LC applications the config.reg file contains the information for the columns. On systems without column tags the column data has to be entered manually.
       * The REG files are rebuild after the next successful start of an instrument session.

    Restart Chemstation & the error could already be fixed.


    2) other steps, if (1) did not help:
    a) take a screenshot of the error
    b) which ChemStation version do you use? (menu: help> about)

     

    copy here additional files:
       c) c:\WINDOWS\ChemStation.ini
       d) instrument LOG file:
                 C:\Chem32\x\TEMP\INSTRx.LOG where X is the instrument number
          e.g. C:\Chem32\1\TEMP\INSTR1.LOG

     

     

    regards, M.