1290 not connecting

Our 1290 was working fine until we moved it a couple of feet to another bench. Now, the modules don’t want to connect, they show yellow lights upon startup.

Going into LabAdvisor, it says “Could not connect to 1290. Ensure that the system is switched on and not connected by another controller”

  • We only disconnected a couple of the CAN cables, and the ethernet to the computer. These are reconnected.
  • The CAN cables are connected to CAN sockets, and I verified they are seated.
  • Ethernet is connected from the DAD to the computer, and is plugged into a LAN connection on the detector, I verified it wasn’t a CAN connector

Please advise what else I should look for.

Thanks in advance!

Parents Reply Children
  • Hi!

     

    Thanks for your reply.

     

    All 6 DIP switches are in the "down" position, however, we didn't touch those while moving the instrument. They look difficult to reset, being recessed. We only disconnected the stack from the computer, disconnected one CAN bus cable so we could move the stack as two parts, and connected it back up. The ping gave the expected negative results, with the DAD by itself. The DAD powers up with a yellow light, as does the rest of the stack, including the column heater. when each individual part of the stack (****, injector, column heater, DAD, A/D converter). The detector is a 1260 Infinity 2

Was this helpful?