Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
GC/MS
  • Technical Areas
GC/MS
Wiki What to Consider When Evaluating Hydrogen User as a Carrier Gas with Agilent EI GC/MS Systems
  • Announcements
  • Blog
  • Forum
  • Articles
  • Files
  • Videos
  • Wiki
  • More
  • Cancel
Actions
  • Share
  • More
  • Cancel
Table of Contents
  • +Collection of GC/MS Resources Home
  • Agilent 5977B Series MSD Ion Source Cleaning and Scheduled Maintenance
  • Agilent 7200 GC/Q-TOF Application Notes
  • Common GC/MS Contaminants and their Source
  • EI filament P/N G7005-60061 replaces both previous EI filaments
  • Explore GC/MS eLearning Courses Available from Agilent Education
  • Extending Gauge Tube Life
  • How to clean an Agilent GCMS Ion Source
  • +How To Get Started with Your New GC or GC/MS System
  • How To Replace a Corrupt Tune File in MSD ChemStation
  • Making a robust 7010 method
  • Minimum Parts Required to Add an MSD to a 6890, 7890 or 8890 GC
  • What to Consider When Evaluating Hydrogen User as a Carrier Gas with Agilent EI GC/MS Systems
  • What to Do If the Helium Carrier Gas to Your GCMS Runs Out
Still Need Help?

Post your question in our User Forum or Contact Support.

What to Consider When Evaluating Hydrogen User as a Carrier Gas with Agilent EI GC/MS Systems

Created by kristen.sensabaugh kristen.sensabaugh over 1 year ago

The purpose of this document is to help users of Agilent EI GC/MS systems determine if hydrogen can be used as a carrier gas for their analyses and what considerations and procedures are necessary to make the transition successful.

EI GC/MS Instrument Helium to Hydrogen Carrier Gas Conversion User Guide (agilent.com)

  • GC/MS
  • Hydrogen carrier gas
  • Share
  • History
  • More
  • Cancel
  • Sign in to reply
Parents
  • ISH
    ISH over 1 year ago

    Helpful info

    • Cancel
    • Vote Up 0 Vote Down
    • Sign in to reply
    • More
    • Cancel
Comment
  • ISH
    ISH over 1 year ago

    Helpful info

    • Cancel
    • Vote Up 0 Vote Down
    • Sign in to reply
    • More
    • Cancel
Children
No Data
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Related
Privacy Statement
Terms of Use
Contact Us
Site Help