Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
GC/MS
  • Technical Areas
GC/MS
Wiki Making a robust 7010 method
  • Announcements
  • Forum
  • Files
  • Videos
  • Wiki
  • More
  • Cancel
  • New
  • +Collection of GC/MS Resources Home
  • Agilent 5977B Series MSD Ion Source Cleaning and Scheduled Maintenance
  • Agilent 7200 GC/Q-TOF Application Notes
  • Common GC/MS Contaminants and their Source
  • EI filament P/N G7005-60061 replaces both previous EI filaments
  • Explore GC/MS eLearning Courses Available from Agilent Education
  • Extending Gauge Tube Life
  • How to clean an Agilent GCMS Ion Source
  • +How To Get Started with Your New GC or GC/MS System
  • How To Replace a Corrupt Tune File in MSD ChemStation
  • Making a robust 7010 method
  • Minimum Parts Required to Add an MSD to a 6890, 7890 or 8890 GC
  • What to Consider When Evaluating Hydrogen User as a Carrier Gas with Agilent EI GC/MS Systems
  • What to Do If the Helium Carrier Gas to Your GCMS Runs Out
Still Need Help?

Post your question in our User Forum or Contact Support.

Making a robust 7010 method

Created by paulsalverda paulsalverda over 1 year ago | Last modified by kristen.sensabaugh kristen.sensabaugh 2 months ago

Simply taking a method from a 7000 series system and attempting to run it under the same conditions on a 7010 series can cause multiple issues:

  1. Saturation - of the source, the quad, and the electron multiplier
  2. Isotope ratio problems
  3. Invalid quantitation results as a result of space charging effects (excessive saturation)
  4. Continual improper use could damage the system!   Don't abuse your GCMS...

 

Here are the steps to figure out how much you can run on a 7010x system:

1.       Run each new sample type in full scan mode (ie MS1Scan) with EM gain set to ~0.5.

If any of the peaks in the scan total ion chromatogram are flat topping (ie overloading the detector), including matrix peaks that you do not need to analyze, please do one or more combinations of the following:

           a.      Inject less until the chromatogram comes on scale.

           b.      Run the sample in split mode (use ultra-inert split liners) or even higher split ratios

           c.      Dilute the sample

           d.      Do further sample clean up

2.       Make sure column flow rate is <1.5 mL/min into the MS while the filament is on

3.       Always run constant flow mode methods. ALL GCMS sources work better under stable vacuum.

4.       Make sure the solvent delay is long enough that all the solvent has eluted before the filament is turned on.

5.       Make sure that there are no leaks.  If you are using any CFT devices, Purged Ultimate Union, splitter, Deans Switch, there are more places leaks can occur.

6.       Always make sure the quad temps are 150C or less. There is no reason to run the quads too hot. They really only need to be hot enough to be stable and eliminate water condensation. Anywhere between 106 and 150 will work up to the maximum source temperature.

7.       Only certain applications need a source temperature >280C, so use 280C as a maximum temperature for the source when running samples.


The 7010 High Efficiency Source system is more sensitive than a 7000.  This may require trade-offs in your methodology.  Use that sensitivity to reduce the amount of sample injected and analyzed - that reduces the amount of normal user maintenance required.   Less liners, gold seals, columns, filaments, and multipliers replaced and less source cleanings save you money and time!

  • GC/MS
  • 7010
  • 7000
  • upgrade
  • Share
  • History
  • More
  • Cancel
Anonymous
  • ISH
    ISH 2 months ago

    Good info

    • Cancel
    • Vote Up 0 Vote Down
    • Reply
    • More
    • Cancel
Related

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Recommended
Privacy Statement
Terms of Use
Contact Us
Site Help