What is the difference between RMSHY-2 (big universal trap for hydrogen) and RMSH-2 (big universal trap for helium)? Can I use helium trap for hydrogen gas? I have teo helium traps and wnat to use for helium and hydrigen gases
What is the difference between RMSHY-2 (big universal trap for hydrogen) and RMSH-2 (big universal trap for helium)? Can I use helium trap for hydrogen gas? I have teo helium traps and wnat to use for helium and hydrigen gases
If you search Agilent.com for the safety data sheets for both: "RMSH-2 SDS" ...and "RMSHY-2 SDS" ..... you'll find that they are both listed on the same one.
Big Universal Trap - Helium-Hydrogen_NAEnglish.pdf (agilent.com)
Having one part number for each type of carrier gas reduces confusion more often.
I would suggest moving to the GasClean filtration system, at least to have one of them installed after the RMSH-2 trap as they have indicators to show when the trap is being used up by bad gas or a leak. When the gas clean indicator starts moving, change the Big Universal Trap ahead of it.
Paul thanks for your reply. I have looked for information about these traps because I have two RMSH-2 traps and I thought about using one for the carrier gas (He) and the other for the FID hydrogen. Anyway, I'm going to buy the hydrogen trap. Thanks again
Paul thanks for your reply. I have looked for information about these traps because I have two RMSH-2 traps and I thought about using one for the carrier gas (He) and the other for the FID hydrogen. Anyway, I'm going to buy the hydrogen trap. Thanks again