While using GC Sampler 80 for headspace analysis, is there a way to delay the split flow so that the split won't be turned on during sample incubation?

We have a 7890A coupled with GC Sampler 80. I noticed that when we use a split method, the split flow will be turned on in the beginning of the incubation. This is quite a waste of helium gas in my opinion. I have tried to look for any delay settings and was not able to find any. Am I missing anything? Is there any way to save helium?

Parents
  • How high a split flow are you using?    Is Gas Saver turned ON in your method?  --like Gas Saver ON, 20 ml/min at 2.00 minutes.      Make sure not to go lower than 20 ml/min as there is a chance of back diffusion of air into the inlet any lower. If you run an oven temperature profile that goes hotter than 250° C, I'd even say that the gas saver should be set to 30 ml/min or even higher to avoid any possibility of oxygen damage to the column.

    What CTC cycle are you using?  Does it list A-xxxx or is it a non-Agilent CTC cycle? 

  • Hi Paul, thank you for your response! the method uses 100:1 split. (I know this is high.... We can try reduce it to 10 or 20). Gas saver is on at 20mL/min at 10 min. I like the info about the oven temp - gas saver flow. Ours goes up to 270 degrees. I'm not sure what CTC cycle mean. The autosampler we are using has these info: CTC Analytics AG, Part# MN01-02; Agilent G6501B.

Reply
  • Hi Paul, thank you for your response! the method uses 100:1 split. (I know this is high.... We can try reduce it to 10 or 20). Gas saver is on at 20mL/min at 10 min. I like the info about the oven temp - gas saver flow. Ours goes up to 270 degrees. I'm not sure what CTC cycle mean. The autosampler we are using has these info: CTC Analytics AG, Part# MN01-02; Agilent G6501B.

Children
Was this helpful?