How to upgrade MS source to Extractor 350?

Hi,

I have a 5977 MS with SST source. I'm thinking of upgrading to Xtr 350. The MS can support it (There's an extra jack on the board to connect the extractor lens), but I was wondering if I need to make changes to the software or run the Parts Finder after installing the source. Any help would be greatly appreciated!

Thank you!

Ehsan

  • G7080B   5977B with Diffusion pump and SS source

    G7081B   5977B with Turbomolecular pump and SS source

    G7077B   5977B with Turbomolecular pump and Extractor source

    G7078B   5977B with Turbomolecular pump, Extractor source, and Chemical Ionization capability

    G7079B   5977B with Turbomolecular pump and High Efficiency Ion Source

    Some parts, like the Ceramic Source Connector board, are used in all of the systems whether the pump, source, electronics, or software support all of the capabilities or not. The diagram is for all models.  The G7080B version of the 5977B only supports the Stainless Steel and Inert EI ion sources.  

  • Thank you for the information, Paul.

    Yes, I saw that information on Agilent manuals and supposed those are just default setups for each equipment and can be upgraded.

  • A G7081B, G7077B, and G7079B can have CI added.  There is not an SS to Extractor upgrade.

    What makes you feel that you need an extractor ion source?  It increases the ion transmission into the quadrupole but most analyses have peaks that are too big already.  There are chromatographic ways to improve overall system signal to noise and that theoretical word "sensitivity."    What are your peaks of interest and at what concentration/amounts?

  • I really appreciate your response, Paul!

    I'm looking to measure hormone concentration in plant tissue using Methyl Chloroformate derivatization. I successfully detected and measured amino acids and some hormones with good sensitivity, but other hormones and some toxins are present in trace amounts (1-100 ng/g fresh weight, very variable depending on species and environment). Sinch most of these compounds have low molecular mass (C5-C10), their spectra don't make some big peaks like diterpenes (C20). I know they are usually being measured using UPLC/HPLC, but we don't have access to this equipment here. It's also very labor intensive and dangerous to use more than 200mg of tissue sample for extraction and analysis using MCF derivatization. Things I'm thinking of is to increase He pressure to 15-16 psi (and put inlet on constant pressure instead of constant flow), so that I can inject more sample without having backflush problem, use SIM instead of TIC, and in the best case, upgrade to Extractor source.

  • Calculate how much of each peak is being injected.

    Don't increase the column flow higher than 1.2 to 1.3 ml/min, but certainly do constant flow.

    How much are you injecting? What solvent is being injected? What are your other injection port parameters?  Split? Splitless  -  what splitless On time?  What splitless purge flow?  What is your initial oven temperature?   You can upload the acqmeth.txt file that exists underneath your method's ******.M subdirectory here.

  • I couldn't upload the file, so I copied the whole text here:

    INSTRUMENT CONTROL PARAMETERS: GCMS
    --------------------------------------

    D:\EhsanKhedive\GC_AquisitionMethod\Total_Metabolome.M
    Mon Mar 20 09:53:06 2023

    Control Information
    ------- -----------

    Sample Inlet : GC
    Injection Source : GC ALS
    Injection Location: Front
    Mass Spectrometer : Enabled


    No Sample Prep method has been assigned to this method.


    GC
    GC Summary
    Run Time 36.357 min
    Post Run Time 1 min

    Oven
    Temperature
    Setpoint On
    (Initial) 45 °C
    Hold Time 1 min
    Post Run 45 °C
    Program
    #1 Rate 7 °C/min
    #1 Value 240 °C
    #1 Hold Time 1 min
    #2 Rate 20 °C/min
    #2 Value 270 °C
    #2 Hold Time 5 min


    Equilibration Time 0.5 min
    Max Temperature 325 °C
    Maximum Temperature Override Disabled
    Slow Fan Disabled

    ALS
    Front Injector
    Syringe Size 10 μL
    Injection Volume 2 μL
    Solvent A Washes (PreInj) 2
    Solvent A Washes (PostInj) 2
    Solvent A Volume 8 μL
    Solvent B Washes (PreInj) 2
    Solvent B Washes (PostInj) 2
    Solvent B Volume 8 μL
    Sample Washes 0
    Sample Wash Volume 2 μL
    Sample Pumps 3
    Dwell Time (PreInj) 0 min
    Dwell Time (PostInj) 0 min
    Solvent Wash Draw Speed 300 μL/min
    Solvent Wash Dispense Speed 6000 μL/min
    Sample Wash Draw Speed 300 μL/min
    Sample Wash Dispense Speed 6000 μL/min
    Injection Dispense Speed 6000 μL/min
    Viscosity Delay 0 sec
    Sample Depth Disabled
    Injection Type Standard
    L1 Airgap 0.2 μL
    Solvent Wash Mode A, B


    Front SSZ Inlet He
    Mode Splitless
    Heater On 260 °C
    Pressure On 7.2892 psi
    Gas Saver Off
    Purge Flow to Split Vent 3 mL/min at 0.75 min

    Thermal Aux 1 (MSD Transfer Line)
    Temperature
    Setpoint On
    (Initial) 280 °C
    Post Run 0 °C


    Column
    Column #1
    Flow
    Setpoint On
    (Initial) 1 mL/min
    Post Run 1 mL/min

    Agilent 19091S-433UI
    HP-5ms Ultra Inert
    -60 °C—325 °C (325 °C): 30 m x 250 μm x 0.25 μm
    In Front SSZ Inlet He
    Out MSD
    (Initial) 45 °C
    Pressure 7.2892 psi
    Flow 1 mL/min
    Average Velocity 36.473 cm/sec
    Holdup Time 1.3709 min

    Column Outlet Pressure 0 psi

    Signals
    Signal #1: Test Plot
    Description Test Plot
    Details
    Save Off
    Data Rate 50 Hz
    Dual Injection Assignment Front Sample

    Signal #2: Test Plot
    Description Test Plot
    Details
    Save Off
    Data Rate 50 Hz
    Dual Injection Assignment Back Sample

    Signal #3: Test Plot
    Description Test Plot
    Details
    Save Off
    Data Rate 50 Hz
    Dual Injection Assignment Back Sample

    Signal #4: Test Plot
    Description Test Plot
    Details
    Save Off
    Data Rate 50 Hz
    Dual Injection Assignment Back Sample


    MS Information
    -- -----------


    General Information
    ------- -----------


    Acquisition Mode : Scan
    Solvent Delay (minutes) : 4.3
    Tune file : D:\MassHunter\GCMS\1\5977\ATUNE.U
    EM Setting mode Gain : 1.000000

    Normal or Fast Scanning : Normal Scanning
    Trace Ion Detection : Off
    Run Time (if MS only) : 650 minutes

    [Scan Parameters]
    Start Time : 4.3
    Low Mass : 31
    High Mass : 400
    Threshold : 150
    A/D Samples: : 4


    [MSZones]

    MS Source : 230 C maximum 250 C
    MS Quad : 150 C maximum 200 C

    Timed Events
    ----- ------
    Number Events= 0

    END OF MS ACQUISTION PARAMETERS


    TUNE PARAMETERS for SN: US1638P001
    ---------------------------------

    Trace Ion Detection is OFF.

    34.593 : EMISSION
    70.007 : ENERGY
    32.907 : REPELLER
    90.331 : IONFOCUS
    17.627 : ENTRANCE_LENS
    1271.928 : EMVOLTS
    1394.3 : Actual EMV
    1.00 : GAIN FACTOR
    2834.000 : AMUGAIN
    138.500 : AMUOFFSET
    1.000 : FILAMENT
    0.000 : DCPOLARITY
    15.482 : ENTLENSOFFSET
    0.000 : Ion_Body
    0.000 : EXTLENS
    323.000 : MASSGAIN
    -25.000 : MASSOFFSET

    END OF TUNE PARAMETERS
    ----------------------

    END OF INSTRUMENT CONTROL PARAMETERS
    ------------------------------------

Was this helpful?