Hello,
Can I use the solid sample PFTBA with the PN: 8500-0656 for the GCMS tuning? If so, pls let me have a procedure to prepare the solution. Thanks.
Hello,
Can I use the solid sample PFTBA with the PN: 8500-0656 for the GCMS tuning? If so, pls let me have a procedure to prepare the solution. Thanks.
PFTBA is a liquid at room temperature. The 8500-0656 is 10grams of liquid - 5.32 ml. 110 µL of PFTBA into an empty vial is safely below the opening of the tube as shown here.
That means that the 5.32 ml (10g) ampule 8500-0656 will fill the calibration vial 48.36 times. Save all the rest in a very tightly sealed vial as it will last many years.
8500-0656 is $304 USD. You could also purchase 05971-60571 which is an ampule that contains 0.5 ml for $124 USD.