After the tune, the EMV is >2400 on our 5975C MS, which I think is too high, is the only solution an exchange or are there other approaches? The built-in EMV is already a few years old and the MS is almost running 24/7.
After the tune, the EMV is >2400 on our 5975C MS, which I think is too high, is the only solution an exchange or are there other approaches? The built-in EMV is already a few years old and the MS is almost running 24/7.
Generally, EM limit is 3000 but some people change it on 2500 based on their application..
How is tune evaluation report ? Is tune evaluation report passed ?
How old is repeller n ent lense of source? Replaced any time or not..
The evaluation report is fine, everything in target range.
As far as I can see the parts have not been replaced since 2012.
The evaluation report is fine, everything in target range.
As far as I can see the parts have not been replaced since 2012.