5973 was running fine a week ago. Now it's giving "Tune Action Stopped: Gain and offset = 0 and peaks are too narrow!" error. Restored an old tune with no luck and also reconfigured system in ChemStation with no luck. Any leads to chase?
Find maintenance, best practice, and troubleshooting articles.
5973 was running fine a week ago. Now it's giving "Tune Action Stopped: Gain and offset = 0 and peaks are too narrow!" error. Restored an old tune with no luck and also reconfigured system in ChemStation with no luck. Any leads to chase?
Thanks Paul. PFTBA vial full, valve audibly clicks and I swapped out to another valve just to be sure. Filaments may be suspect but visually look ok and show continuity with multi meter. Vacuum with column installed is 1x10-5. Source cleaned couple months ago but very little sample load. Pump oil only couple months old. I blanked the transfer line taking the GC out of the equation, still same error.
Hi MarkH
I don't think it's side board or HED power supply problem.
We faced same problem.
One time column was more out from transferline in MS.
2nd time, We cleaned source ( replaced repeller as it was very scratchy and replaced EM horn and problem was solved.
Generally, we are keeping one extra Ion source assy and EM Horn in Lab to verify the problem.
Generally, we are keeping one extra Ion source assy and EM Horn in Lab to verify the problem.