I have an HP 5973 MSD. When I tried to tune it this morning this error messaged popped up " tune action stopped: gain and offset = 0 and peaks are too narrow. Any suggestions on how to fix it?
Find maintenance, best practice, and troubleshooting articles.
I have an HP 5973 MSD. When I tried to tune it this morning this error messaged popped up " tune action stopped: gain and offset = 0 and peaks are too narrow. Any suggestions on how to fix it?
Hi,
A good place to start is to evaluate the last couple of tunes and see if you notice any trends. High repeller, air leaks etc. In a lot of cases a source cleaning, or tuning from defaults can resolve this type of tuning issue.
Regards
James
Hi,
A good place to start is to evaluate the last couple of tunes and see if you notice any trends. High repeller, air leaks etc. In a lot of cases a source cleaning, or tuning from defaults can resolve this type of tuning issue.
Regards
James