I have an HP 5973 MSD. When I tried to tune it this morning this error messaged popped up " tune action stopped: gain and offset = 0 and peaks are too narrow. Any suggestions on how to fix it?
I have an HP 5973 MSD. When I tried to tune it this morning this error messaged popped up " tune action stopped: gain and offset = 0 and peaks are too narrow. Any suggestions on how to fix it?
Hi,
When ion source cleaning performed last time?
Did you change column?
The ion source was cleaned in May 2020. The column was changed in January 2020
Check column length at outlet of transferline.
Check how much Volt is showing for EM Horn in your last autotune if u have print out.
How is fliament..
Pump oil level?
Hi,
A good place to start is to evaluate the last couple of tunes and see if you notice any trends. High repeller, air leaks etc. In a lot of cases a source cleaning, or tuning from defaults can resolve this type of tuning issue.
Regards
James