7890A/5975C/7697A Headspace-GC-MS slow loading of software

Hi everyone,

 

We have a 7890A with mass spec (5975C) and head space analyzer (7697A) running MassHunter acquisition (B.07.04.2260) on a Windows 7-64 bit computer.

The headspace control software is B.01.06, the GC drivers package is A.03.02.021

 

Whenever we load the software there is huge lag (up to 1 minute) in the software being run and the software being ready for use. During the software loading, the Windows activity monitor shows on two separate occasions that the program is not responding. Not responding is shown for at least several seconds each time. It seems to me that the end of the not responding period is accompanied both times by an audible beep from the headspace.

This issue has persisted through more than one version of the MassHunter software. I can't remember if we had the problem with MSD Chemstation or not

 

We have talked to Agilent about this and have gone through the normal steps - reinstall the firmware, remove the software and reinstall. reinstall Windows and reinstall windows and then the software. I even tried installed Windows from an installation disk on a brand new PC and then installing the software but I had installed anything else. No success.

 

 

Any ideas?

Thanks in advance

Kevin

Parents
  • One of my co-workers just pointed me in the direction of something I didn't know of.

    This is a Windows 'feature' where the system is trying to go out on the Internet and verify that the software certificate is still valid even if the system is not connected.

     

    Tools->internet options->advanced tab->under security uncheck  “check for publisher’s certificate revocation”.

     

Reply
  • One of my co-workers just pointed me in the direction of something I didn't know of.

    This is a Windows 'feature' where the system is trying to go out on the Internet and verify that the software certificate is still valid even if the system is not connected.

     

    Tools->internet options->advanced tab->under security uncheck  “check for publisher’s certificate revocation”.

     