After cleaning the source and pumping down the 5973 will not reload and gives the 24 and 26 error codes. I have to reboot the computer to fix the problem.
Find maintenance, best practice, and troubleshooting articles.
After cleaning the source and pumping down the 5973 will not reload and gives the 24 and 26 error codes. I have to reboot the computer to fix the problem.
Does it state anything with the error messages? Like "source cannot reach temperature" or anything like that?
If this wasn't happening prior to cleaning the source but it happening now after cleaning the source. I would vent tge MSD, turn it off when it's vented, take the source out, disassemble it, clean it with your choice of solvent, let it dry at an appropriate temperature, reassemble it and then put it back into the MSD.
Ensure all the wires are fitted to the correct positions and pump down the system. Once the turbo pump reaches 100% speed, turn on the software 9chemstation/masshunter/etc) and let it reach temperature. Then let it settle for two hours to stabilize.
Does it state anything with the error messages? Like "source cannot reach temperature" or anything like that?
If this wasn't happening prior to cleaning the source but it happening now after cleaning the source. I would vent tge MSD, turn it off when it's vented, take the source out, disassemble it, clean it with your choice of solvent, let it dry at an appropriate temperature, reassemble it and then put it back into the MSD.
Ensure all the wires are fitted to the correct positions and pump down the system. Once the turbo pump reaches 100% speed, turn on the software 9chemstation/masshunter/etc) and let it reach temperature. Then let it settle for two hours to stabilize.