Difficulty with the HED supply "error, MS fault 2  2

Am having the above error message on agilent 5975 MSD whiles checking for Air and water. The process cannot continue and had to shutdown the instrument without venting. Though I suspect the HED power supply, I would like to have some confirmation before moving forward.

Parents
  • Hi,

     

    this error message could happen if there is a really huge amount of air/water in your system. If the error occurs after pump down, you should wait about 1-2 hours to move forward.

    Otherwise it could be a problem with the power supply itself or the HED.

    Try to clean the HED, if this doesn't help call the Service to test/replace the power supply.

     

    BR

    PWD

Reply
  • Hi,

     

    this error message could happen if there is a really huge amount of air/water in your system. If the error occurs after pump down, you should wait about 1-2 hours to move forward.

    Otherwise it could be a problem with the power supply itself or the HED.

    Try to clean the HED, if this doesn't help call the Service to test/replace the power supply.

     

    BR

    PWD