I made the mistake of doing a standard tune multiple times instead of an auto-tune and I know this lowers the sensitivity according to the hardware manual. Could this be the source? If so, how can I fix it?
I made the mistake of doing a standard tune multiple times instead of an auto-tune and I know this lowers the sensitivity according to the hardware manual. Could this be the source? If so, how can I fix it?
Hello xio,
the case is a bit older: do the problems still exist?
If so, a new tune report would be more than helpful.
small hint to the first tune report (24th July):
a) If the H20 value is too high (e.g. like in the first 1-2h after pump down: the H20 value is gigantic) then usually the N2/O2 is also very high, WITHOUT a leak!
Therefore: if the H2O value is higher than 5-10%, first look for the cause of the water reason.
Pump down: after one night the H2O value is usually less than 3(-5)% if the MSD is supplied with column flow.
If the MSD is NOT supplied with carrier gas (blind sealed) the MSD needs much longer to pump down the H2O from the MSD; here I have seen several days higher H2O values.
b) Number of peaks=512
this value is also MUCH too high.
it could also drop to a normal level 100-150 peaks if the H2O value is at a normal level
regards, M.
Just wanted to provide a long overdue update. The issue was caused by a bad circuit board.