Hello,
I have a HP 6890 Plus that I need to change the IP address on. Can anyone guide me on how to do that? It has been a while since I've done it. Do I connect my PC to it and change the IP through the network settings?
Thanks,
-Lara
Hello,
I have a HP 6890 Plus that I need to change the IP address on. Can anyone guide me on how to do that? It has been a while since I've done it. Do I connect my PC to it and change the IP through the network settings?
Thanks,
-Lara
Hi Lara,
The "G1369C LAN Interface manual" will cover how to configure the LAN card. A google search will point you to the manual.
In short
If your PC is connected to the same network as the GC
Open a command prompt on PC
Telnet <Current Address of GC>
ip <New Address for GC>
quit
Power Cycle GC
FYI, It might be more complicated depending on your network, PC and current GC configuration. If you need more step by step I can go into more detail.
Tim
The manual can be found at https://www.agilent.com/cs/library/usermanuals/public/G1369C_LAN-InterfaceCard.pdf. Look at the section on setting the configuration switches on pages 21-26. Instructions for using Telnet start on page 39.
Telnet may need to be enabled on the PC. To do that, open Windows Control Panel > Programs and Features. Click on "Turn Windows features on or off" and a new window will open listing the Windows features. Select "Telnet Client" and click OK. Admin access and permissions may be required to complete this.