GC 7890B septum purged remains 0.1 while it is set to 3.0 mL/min.

GC 7890B purge back inlet septum purged remains 0.1 while it is set to 3.0 mL/min. 

What could be the possible solutions for this problem?

Parents
  • HiCarlosDav

    It seems, the outlet line of septum purg closed by liquids. Or the line closed in the end of line. Please check it.

    And in acquisition software, you cal increase the flow of septum purge from 3 to 6 or opper to push the liquid for going out. Or by a hair heater warm the line for many times to going out the liquids.

Reply
  • HiCarlosDav

    It seems, the outlet line of septum purg closed by liquids. Or the line closed in the end of line. Please check it.

    And in acquisition software, you cal increase the flow of septum purge from 3 to 6 or opper to push the liquid for going out. Or by a hair heater warm the line for many times to going out the liquids.