Problem with PPU column

I am using MicroGC-490. Channel 1 has Mol. sieve 5Å column and channel 2 has PPU column. The GC has be shut for few months. Now, the signal from PPU column is acting strangely. The Mol. sieve channel is working properly. Can some help me identify what might have caused the problem. Attached are the detector response now  and before.

Parents
  • Hello  ,

    I am rather unfamiliar with the Micro GCs, but just looking at the two chromatograms, it looks the signal goes from 3.3 mV approximatively to 0 mV, then stays there. It is hard to tell on the original chromatogram where the baseline is at, but if you can zoom in, I'd be curious to see. Is it at 3.3, is it at 0?

    What type of detector are you using? Only asking because I am assuming it could be a detector problem, not a column problem.

    Hopefully someone with knowledge on Micro GCs will be able to provide a better answer on this.

  • Hi

    The detector is microTCD. Both the signals are from running ambient air sample with exactly the same settings. Now, there is slight response in the beginning and then it goes down to zero as you noted. When it was functioning normally, there was a composite peak (everything in the air) at 0.348s followed by a CO2 peak at 0.401s. These signals are completely missing now.

    I don't have an extensive knowledge on GC systems in general. I have used it only occasionally. So, I was wondering if someone has some insights into this problem.

Reply
  • Hi

    The detector is microTCD. Both the signals are from running ambient air sample with exactly the same settings. Now, there is slight response in the beginning and then it goes down to zero as you noted. When it was functioning normally, there was a composite peak (everything in the air) at 0.348s followed by a CO2 peak at 0.401s. These signals are completely missing now.

    I don't have an extensive knowledge on GC systems in general. I have used it only occasionally. So, I was wondering if someone has some insights into this problem.

Children
Was this helpful?