Vial pickup on Tray on 7683 autosampler frozen

Hi everyone

The vial pickup on our autosampler tray arm stopped working during the course of a series of injections and appears to be locked. The arm still rotates but the pickup is locked.

We have tried the obvious - switch it on and off but without resolution

Any suggestions. I think that we had this problem many years ago and there was a solution but I can't remember what it was

Thanks in advance

Kevin

Parents Reply
  • If the up and down won't freely move through the full range of motion and locks up at a certain point the arm is toast. You can order a replacement arm (G2614-60531). If you have the budget I would recommend buying a newer ALS.

    If you want to take it apart and tinker with it remove all the trays and look underneath the arm, there's a Phillips head screw on each end. Take care not to move the rods the arm rides on from the position they were in after disassembly.