071619
Hi,
Which calculation method is better in Chemstation External standard calculation or Internal standard calculation? Is there a way to utilize both for better accuracy and precision? Any relevant response is welcome. Thanks!
Find maintenance, best practice, and troubleshooting articles.
071619
Hi,
Which calculation method is better in Chemstation External standard calculation or Internal standard calculation? Is there a way to utilize both for better accuracy and precision? Any relevant response is welcome. Thanks!
Hi megaflorch,
First of, a bit of explanation.
External calibration:
You have got one ore more standards with known concentration. You are using these to connect the measured response (mainly peak area) to the amount of substance in your standard. The resulting calibration is used to calculate the amount of substance in an unknown sample.
Internal calibration:
You have got one ore more standards with known concentration. You are using these to connect the measured response (mainly peak area) to the amount of substance in your standard. The resulting calibration is used to calculate the amount of substance in an unknown sample.
ADDITIONALLY you have got a substance in all standards, blanks and samples, preferably at the same concentration.
The substance must not be part of your sample, but should be as similar as possible. In high end MS, deuterated substances are often used.
This substance is going to be added before sample preparation. If you are losing some of your substance due to sample preparation you have lost some of the internal standard as well and you can use this known loss to compensate the other unknown amounts in calculation.
So, Internal Calibration essentially already includes external calibration.
Internal Calibration usually give you better results if you sample undergoes extensive preparation steps with a lot of possibilities to lose substance.
Prerequisites are that you do have a substance you can use as internal standard.
I hope this helps.
Best regards,
Dominik
Hi megaflorch,
First of, a bit of explanation.
External calibration:
You have got one ore more standards with known concentration. You are using these to connect the measured response (mainly peak area) to the amount of substance in your standard. The resulting calibration is used to calculate the amount of substance in an unknown sample.
Internal calibration:
You have got one ore more standards with known concentration. You are using these to connect the measured response (mainly peak area) to the amount of substance in your standard. The resulting calibration is used to calculate the amount of substance in an unknown sample.
ADDITIONALLY you have got a substance in all standards, blanks and samples, preferably at the same concentration.
The substance must not be part of your sample, but should be as similar as possible. In high end MS, deuterated substances are often used.
This substance is going to be added before sample preparation. If you are losing some of your substance due to sample preparation you have lost some of the internal standard as well and you can use this known loss to compensate the other unknown amounts in calculation.
So, Internal Calibration essentially already includes external calibration.
Internal Calibration usually give you better results if you sample undergoes extensive preparation steps with a lot of possibilities to lose substance.
Prerequisites are that you do have a substance you can use as internal standard.
I hope this helps.
Best regards,
Dominik