Hello, how can you configure an Agilent 7890B GC System w/7697A Headspace Sampler to do liquid injections with an ALS tower? The headspace transfer line is integrated into the GC inlet gas line.

I have created a new method to perform liquid injections with the ALS. The inlet pressure will not stabilize and never becomes ready for injection. How should the inlet carrier flow be plumbed for liquid injections considering that there is an integrated headspace line attached to the inlet port.

Thank you for your help,

Lisa

Parents Reply
  • Actually, now, as you have removed heaspace then check carrier line of inlet from EPC..Is it via headspace or direct from EPC.

    But before that replace septa and tight nut properly and check for leak from column connection in oven.

    And if you have maintenance experience then you can check all connection of carrier going to GC inlet.

    And last, you can call to your local agilent support to do modification to put require setup onsite...