Contamination/unwanted Peaks polar Column on GC-FID

Hi Everyone

This is my first post here in this community. And I really hope to get some ideas to fix my problem, since i'm getting slowly desperate!

So here's my problem: We have some really bad contamination/peaks on all our polar Systems: 

I especially talk about the peaks after 5min. it occurs on both systems, and the big broad peak appears also on both systems (3 colums in total). We have an Agilent GC 6890N and a 7890B (Dual, two colums) with the same colums installed: DB-WaX-FF 20m/01mm/0.2µm. All have the same oven method. We always use split-methods.

Issues appeared suddenly after a liner-change. But we use those liners on other systems as well and there we have no problems. I changed column, gold-plate, septum, liner, cleaned the detector an syringe and changed the FID-Jet and even the solvent trap as well. After all this, it looked more or less the same. I suspected a contamination from outside (solvent) but when i run instrument blanks, i can see the same peak contamination, so it must be something inside the system.

We have a Standard-solution we ran every day, and with this we can monitor our devices. on one day, (this monday) none of all the peaks on both systems appeared in the chromatogram: (ignore the two peaks at the beginning, it comes from the solvent MTBE)

but after this day, peaks were back as the day before.

I need to add: in the middle of last month, we had a problem with H2 quality (oxygen/water in the system) that lead to high column bleed and the columns were strongly damaged, that we needed to change all of them (That was this coulmn change i mentioned before). But nevertheless, the peaks i showed on the grafics were on the columns before and after that.

I hope i did not miss any important information. I am very happy to hear your ideas on this problem.

Thank you very much in advance!

KR 

Parents Reply Children
  • How old is column?

    Cut little length of column from inlet and detector side...

  • All Columns were changed shortly after the problem occured. But I will try it with shortening the column again and I will see again. thanks!

  • Hi

    The Problem is solved! So we didn't shortened the column it, since instrument blanks were alright. The Problem was actually quite "easy" to solve and had nothing to do with the column.

    Here the solution (Maybe helpful for other user)

    We used to store our solvent in Vials in the fridge, since we use them at the beginning of every week. After the time we started to prepare a few of them in advance, that we don't have to prepare the vials every week from the start. But that also lead to this problem. It seems like, after a while, the septum in the lid of the vials started to diffuse into the solvent. That caused extra peaks that accumulated during the week:

    (Same Vial, measured every morning, monday to friday)

    Now, with totally fresh filled vials, all systems are clear.

    Now everything makes sense, pretty obvious but a long way to solve it!

    Thanks again for the help!

Was this helpful?