Any help to resolve this error.
Restarting done.
Plunger replaced.
Syringe replaced.
But still this error coming after completing 2 or 3 Run...
Find maintenance, best practice, and troubleshooting articles.
Any help to resolve this error.
Restarting done.
Plunger replaced.
Syringe replaced.
But still this error coming after completing 2 or 3 Run...
From the 7693 ALS manual:
Perhaps try adding (or extending) viscosity delay to see if that helps. Otherwise it may need to be sent to Agilent Service Center for repair.
Thanks for reply.
Repairing done...
Thanks for reply.
Repairing done...