Hello, FID Signal output is showing 576675 in 7890b gc..Even after removing FID Block assy, same signal is showing..It is not zeroing...What could be reason of this problem?
Hello, FID Signal output is showing 576675 in 7890b gc..Even after removing FID Block assy, same signal is showing..It is not zeroing...What could be reason of this problem?
Hi,
could be a short of the electrometer and/or the spring. try to uninstall/clean the spring and check again. Could also be a problem on the signal board. If the error persist i suggest to call Agilent service to replace/check electrometer/signal board.
BR
PWD
Thank you for reply...May i know that electrometer of 6890N and 7890B is same or not...
Thank you for reply...May i know that electrometer of 6890N and 7890B is same or not...