FID SIGNAL PROBLEM

Hello, FID Signal output is showing 576675 in 7890b gc..Even after removing FID Block assy, same signal is showing..It is not zeroing...What could be reason of this problem?

Parents
  • Hi,

    could be a short of the electrometer and/or the spring. try to uninstall/clean the spring and check again. Could also be a problem on the signal board. If the error persist i suggest to call Agilent service to replace/check electrometer/signal board.

    BR

    PWD

Reply
  • Hi,

    could be a short of the electrometer and/or the spring. try to uninstall/clean the spring and check again. Could also be a problem on the signal board. If the error persist i suggest to call Agilent service to replace/check electrometer/signal board.

    BR

    PWD