Designing the ideal standby method

Hello all

 

I'm looking at designing some standby methods for my 4 GC-FIDs and 3 GC-MSs as they are often sitting idle for 3/4 days a week and are happily consuming helium, hydrogen and generally sucking money away basically for no reason. Turning them off/on is not an option as we often need some of them fairly urgently, so I thought a 'Standby' method would be the best way to conserve!

 

I just want some advice on the key things to be looking at regulating. From scrawling this and other forums I've pieced together the following information so if anybody spots any glaring holes, please let me know

 

Oven temperature: Low (30-50)

Column: Carrier gases set to a minimum flow (~0.5ml?) to protect column and/or MS (we are an oils lab so use non-polar columns)

FID: Keep at around 100 to prevent moisture forming on jet. Turn off all gases except for carrier gas

MS: Keep at temperature to ensure no build up of moisture/solvent anything on source etc. Same with aux line?

 

My main point of concern is the inlet. We have split and splitless injectors - how much flow do we need to keep going for the split flow? Any? Low? I've seen some people say to set the method to splitless and put a low flow on it (1-2ml/min). Currently it's on Gas Saver running 20ml/min which I feel is fairly high compared to the remainder of the system.

 

Any advice on this will be appreciated!

Parents
  • Have you considered gas generators for Air, H2 and N2?

    Helium could be substituted by H2 as a carrier and, from what I read, this allows for faster analysis.

    The danger is H2 accumulation in oven (broken column) or MSD (stopped pump).

     

    What do you think ?

    Would there be much saving this way?

  • Hi csiquet_fles2000,

    Generators are a viable solution as long as they are well maintained. The switch on the carrier gas side is not straight forward as there are method and column changes that will be needed in most cases. One reference is below:

    Converting EI GC/MSD to Hydrogen 

     

    For anyone considering the change to H2, it is always good to reach out to your local support and review your instrument configuration to make sure the change is applicable to your hardware. For example H2 would not be appropriate with a purge and trap system or headspace. There are alternate solutions in each case. Agilent contact centers can provide information on applications as well as safety information.

    Regards

    James

  • Thanks James,

    I'm undecided as to what to do - as it is just a short shutdown (I guess I'm just concerned if there was a power out during the shutdown as no-one will be onsite to alert us to it).

    Regarding the set-up of the instrument - can you advise if labs generally exhaust the split vents to fume cupboards/extraction as a rule? The site of the GCMS is not close enough to external walls for outside extraction so currently the vents are just into the lab (I've attached a pic). And finally - I've seen recommendations for changing split traps - can you advise on an approximate schedule for this (I realise it will somewhat be affected by number of samples/type of samples - but an approximate would just help in the set-up of maintenance scheduling). Thanks! Louise

  • Hi louh,

    For the most part with split vent , its not common, but is a good idea. I have had labs do this for high H2 splits or solvent types such as Cs2. The limiting factor, as you mentioned, is that the instruments are generally not close to a fume hood. There are some nice set ups with ventilation trunks. Image from a quick web search as an example below. Laboratory air quality is always a concern, so anything that can improve quality is a good idea.

    For split vent trap changes, you will actually become the expert on your system if you are not having AGilent scheduled Preventative Maintenance visits. You are correct in that sample type and the amount of samples run play a factor in replacement. On an annual PM the trap should be changed. For some customers this can be too long , FAME application for example with high throughput, for other applications a year is fine.

     

    Regards

    James

  • Hi James,

    I'm wondering if you can help with a Chemstation G1701 software question? I noticed in the Agilent OQ info online a mass ratio precision can be done during the injection precision test. I've done the precision test and just wondering how I can get the software to give me a ratio of 2 ion masses compared for each injection - I've attached what I think I need to achieve but have yet to find how to do this in any of the software manuals or Help topics! (As previously mentioned this instrument was bought second hand and I found this file in the files put on there by the vendor)

    Thanks!

    Louise

  • Hi James,

    I'd like to ask: how critical is the capping of the ends of the columns?

    I know columns should be flushed with carrier gas before heating the oven but, when stored at room temperature, is it likely to get damaged if left uncapped?

    Capping is not always possible. For example, we have one column end with a CFT nut and it is not possible to cap with a septum.

    As well, I have read that the He of a capped column would slowly diffuse out and create a vacuum. Wouldn't air diffuse in then?

    Thanks.

  • Hi csiquet_flex2000,

    Capping for storage is critical to minimize the column phases exposure to oxygen even at room temperature. I haven't heard of He diffusing  through fused silica and creating a vacuum. This is possible with plastic type tubings (how a LC degaser works). I have seen uncapped and capped columns in the same labs before, capping with septa or column end caps do make a difference in column health. 

    Regards

    James

Reply
  • Hi csiquet_flex2000,

    Capping for storage is critical to minimize the column phases exposure to oxygen even at room temperature. I haven't heard of He diffusing  through fused silica and creating a vacuum. This is possible with plastic type tubings (how a LC degaser works). I have seen uncapped and capped columns in the same labs before, capping with septa or column end caps do make a difference in column health. 