Designing the ideal standby method

Hello all

 

I'm looking at designing some standby methods for my 4 GC-FIDs and 3 GC-MSs as they are often sitting idle for 3/4 days a week and are happily consuming helium, hydrogen and generally sucking money away basically for no reason. Turning them off/on is not an option as we often need some of them fairly urgently, so I thought a 'Standby' method would be the best way to conserve!

 

I just want some advice on the key things to be looking at regulating. From scrawling this and other forums I've pieced together the following information so if anybody spots any glaring holes, please let me know

 

Oven temperature: Low (30-50)

Column: Carrier gases set to a minimum flow (~0.5ml?) to protect column and/or MS (we are an oils lab so use non-polar columns)

FID: Keep at around 100 to prevent moisture forming on jet. Turn off all gases except for carrier gas

MS: Keep at temperature to ensure no build up of moisture/solvent anything on source etc. Same with aux line?

 

My main point of concern is the inlet. We have split and splitless injectors - how much flow do we need to keep going for the split flow? Any? Low? I've seen some people say to set the method to splitless and put a low flow on it (1-2ml/min). Currently it's on Gas Saver running 20ml/min which I feel is fairly high compared to the remainder of the system.

 

Any advice on this will be appreciated!

Parents
  • Have you considered gas generators for Air, H2 and N2?

    Helium could be substituted by H2 as a carrier and, from what I read, this allows for faster analysis.

    The danger is H2 accumulation in oven (broken column) or MSD (stopped pump).

     

    What do you think ?

    Would there be much saving this way?

  • Hi csiquet_fles2000,

    Generators are a viable solution as long as they are well maintained. The switch on the carrier gas side is not straight forward as there are method and column changes that will be needed in most cases. One reference is below:

    Converting EI GC/MSD to Hydrogen 

     

    For anyone considering the change to H2, it is always good to reach out to your local support and review your instrument configuration to make sure the change is applicable to your hardware. For example H2 would not be appropriate with a purge and trap system or headspace. There are alternate solutions in each case. Agilent contact centers can provide information on applications as well as safety information.

    Regards

    James

  • Hi James,

     

         Thank you for replying so swiftly. I am too silly thought that I could use 1 Helium gas cylinder for more than 1 year. As the covid lock down, the lab was shut for a couple of months, the GCMS wasn't use much for the year. The GC consumable and standard could save apart form the Helium. Thank you for the calculation. I might consider Hydrogen generator if possible for my method, a bit of research and method development for the next year. Thank you every one here. I have learnt a lot that couldn't be learnt in any courses or books

     

    Thanks

     

    Candy

  • Hi James,

    I have recently started at a company that has purchased a second hand Agilent 7890A and 5595MSD (G3171A) and I'm in the process of getting it in service and setting 'routine operations'. As it was a second hand purchase we don't have an agilent representative to enquire on technical advice so I'm reading a lot of manuals! I'd like to ask your advice on what to do with the instrument over a christmas shutdown....we shut from 23rd Dec - 4th Jan - should the standby method just be used for the GC parameters (as discussed in this thread) and the MSD be kept at temp or should the MSD be fully vented and shutdown?

    Thanks 

    Louise  

  • Hello

     

    I'm not James but we also closedown over Christmas and always turn off all of our instrumentation for the week.

     

    Vent the MS, shut it down and turn off the GC. Nothing usually goes too wrong, though depending on the column you use might be worth removing it and plugging the ends to prevent things from oxidising. Probably good for the MS to have a bit of a rest - without being stop/started!

  • Hi Louise,

    Some customers shut down like ryanedwardson mentioned. If you shut down and remove the column definitely plug the ends. I would also cap( blank no hole ferrule of the proper type 5181-3308 | Agilent ) the msd interface and start the MSD for about a 30 seconds or so to put some vacuum on the analyzer. I usually prefer that just a standby method be loaded and the instrument left under vacuum with flow on the column, but customers are successful with both scenarios.

    Regards

    James 

  • Thanks James,

    I'm undecided as to what to do - as it is just a short shutdown (I guess I'm just concerned if there was a power out during the shutdown as no-one will be onsite to alert us to it).

    Regarding the set-up of the instrument - can you advise if labs generally exhaust the split vents to fume cupboards/extraction as a rule? The site of the GCMS is not close enough to external walls for outside extraction so currently the vents are just into the lab (I've attached a pic). And finally - I've seen recommendations for changing split traps - can you advise on an approximate schedule for this (I realise it will somewhat be affected by number of samples/type of samples - but an approximate would just help in the set-up of maintenance scheduling). Thanks! Louise

Reply
  • Thanks James,

    I'm undecided as to what to do - as it is just a short shutdown (I guess I'm just concerned if there was a power out during the shutdown as no-one will be onsite to alert us to it).

    Regarding the set-up of the instrument - can you advise if labs generally exhaust the split vents to fume cupboards/extraction as a rule? The site of the GCMS is not close enough to external walls for outside extraction so currently the vents are just into the lab (I've attached a pic). And finally - I've seen recommendations for changing split traps - can you advise on an approximate schedule for this (I realise it will somewhat be affected by number of samples/type of samples - but an approximate would just help in the set-up of maintenance scheduling). Thanks! Louise

Children
  • Hi louh,

    For the most part with split vent , its not common, but is a good idea. I have had labs do this for high H2 splits or solvent types such as Cs2. The limiting factor, as you mentioned, is that the instruments are generally not close to a fume hood. There are some nice set ups with ventilation trunks. Image from a quick web search as an example below. Laboratory air quality is always a concern, so anything that can improve quality is a good idea.

    For split vent trap changes, you will actually become the expert on your system if you are not having AGilent scheduled Preventative Maintenance visits. You are correct in that sample type and the amount of samples run play a factor in replacement. On an annual PM the trap should be changed. For some customers this can be too long , FAME application for example with high throughput, for other applications a year is fine.

     

    Regards

    James

Was this helpful?