Shutdown (#6): Back Inlet Flow Shutdown - Agilent 6890

Hello community,

I set up an Agilent 6890 (Software: ChemStation A08.03).

A few seconds after I start the GC, the error message "Shutdown ( # 6 ): Back Inlet Flow Shutdown" is shown on the display. The settings of the Back Inlet shows that the pressure is set to 0.29 but the actual pressure is 0.00. I already checked the helium inlet to the GC: It is set to 6 bar and the gas pipe are clean and nice. I also changed the septum. But the problem is still present. The inlet is connected to a 30 m x 0.250 mm column and the FID is used as detector.

Has anybody an idea how I can fix this problem?

 

Thank you!

Parents Reply
  • Hello @garyb007,

    tank you very much for your answer!

     

    My teammate has changed the system from standard analysis to headspace some weeks ago. Therefore he used the front linet. Now I connected the column back to the back inlet. I took some photos of the column and the inlet. Maybe you see an error. If you need more pics or information feel free to ask

     

    Best regards

    Alexander

Children
No Data
Was this helpful?