7683 Series ALS on 6890N GC won't inject

I have a 6890N GC-FID with a 7683 Series ALS and I'm using Agilent Open Lab (EZChrom Edition) software.  The software won't trigger the autosampler to make an injection.   A box appears on the screen with the message 'Incompatible Tray'.  I have verified that "External" is chosen as the source of the injection in the software.  I've replaced the autosampler tray with another tray that is known to be working, and I've replaced the tray cable.  I've also rebooted the GC and the computer.   None of these things have corrected the problem.  Is there anything else I can try, or does this indicate a bad board?

Parents Reply
  • Are you able to post a screenshot of the error message? It could be helpful.

     

    By the way, with a firmware revision of A.03.08, the GC is either a 6890A or 6890 Plus, not a 6890N. You have a separate box that is the sampler controller, right? In the 6890N, the injectors and tray plug directly into the back of the GC.

     

    I agree that it doesn't seem likely that there is a hardware problem if the instrument can make a run from the front panel.