Equipment used 7890A natural Gas GC with extended analysis. During operation, FID is frequently shut-off. Try to ignite manually and temporarily solve the problem. But what need to check for identifying the problem?
Equipment used 7890A natural Gas GC with extended analysis. During operation, FID is frequently shut-off. Try to ignite manually and temporarily solve the problem. But what need to check for identifying the problem?
Please see the attached problem statement and method parameters for better understanding. Let me know if additional data you need.
Please see the attached problem statement and method parameters for better understanding. Let me know if additional data you need.
I would still focus on the FID, the PCM shut down and other shut downs could be related to hydrogen safety shut downs. Simple and quick would be to change the jet or in the very least run the test that apcharles suggested.