Equipment used 7890A natural Gas GC with extended analysis. During operation, FID is frequently shut-off. Try to ignite manually and temporarily solve the problem. But what need to check for identifying the problem?
Equipment used 7890A natural Gas GC with extended analysis. During operation, FID is frequently shut-off. Try to ignite manually and temporarily solve the problem. But what need to check for identifying the problem?
I have updated the tags for better visibility of your post. Could you reply with the FID parameters that you are using? Temperature, flows etc? If there hasn't been any problems in the past, the most likely cause is that the jet needs to be changed.
James