6890 GC flow issues FID/TCD

I am currently having a flow issue with my 6890N GC(w/ G1888 HS).  I am doing yearly flow checks.  Hooked up a calibrated flow meter to DET exhaust and turned off all gases (pressure off, makeup He off, Air off everything). I then enabled them one at a time at varied levels.  FID= detector controlled gases worked fine and in spec.  When I set the column flow the reading was low out of spec tried the varying levels still no go.  Switched to the front which is a TCD.  Same result detector controlled gases worked fine, but as soon as I went to do the flow through the column low.  TCD was lower than the FID (TCD=1.36 on set of 2.5, FID was 1.8 on same setting).  Checked gas filters they were old but no problem.  2 different columns but check all parameters and re installed to ensure good connection.  Not really sure where else to go at this point.  Any help or suggestion would be greatly appreciated.

 

THANKS!

 

 

Parents
  • Hi eme091,

    When trying to measure smaller flows with a flow meter there can be some difficulty. The volume of the meter tubing(should be small),adapter to the detector (tight and proper fit) etc can play a factor. Also I have had a meter that had trouble with low flows and it was only resolved by making sure the meter had the gas type set correctly and was attached to the detector output and left alone (not moved)  for 10 minutes to stabilize . There are also some meters out there that are not spec to read low flows. What meter are you using?

    regards

    James

     

    I updated the tags on this post so it will have more visibility in the community and others can comment.

  • It is an agilent brand flow meter, will check the model when I get back to work but it had no problem on another instrument a week or 2 ago and one of the detector readings done was at a similar flow.  I'm starting to wonder if it's the EPC board?

     

    flow meter is Agilent ADM2000 universal gas flowmeter, was just calibrated 31JUL17

  • There could be a problem with the EPC or the EPC board, but are you seeing any chromatographic symptoms? Are peaks eluting later than they should? Are peak areas diminished?