6890 GC flow issues FID/TCD

I am currently having a flow issue with my 6890N GC(w/ G1888 HS).  I am doing yearly flow checks.  Hooked up a calibrated flow meter to DET exhaust and turned off all gases (pressure off, makeup He off, Air off everything). I then enabled them one at a time at varied levels.  FID= detector controlled gases worked fine and in spec.  When I set the column flow the reading was low out of spec tried the varying levels still no go.  Switched to the front which is a TCD.  Same result detector controlled gases worked fine, but as soon as I went to do the flow through the column low.  TCD was lower than the FID (TCD=1.36 on set of 2.5, FID was 1.8 on same setting).  Checked gas filters they were old but no problem.  2 different columns but check all parameters and re installed to ensure good connection.  Not really sure where else to go at this point.  Any help or suggestion would be greatly appreciated.

 

THANKS!

 

 

Parents
  • So the EPC board did fix the back flow, I had improperly zeroed the new board.  So now the back flow is hitting right on the nose.  The front flow is still low.  if I increase the oven temp from 50 to 100 the flow reading moves from 1.4 to 1.78 on the meter (which I did verify on another instrument).  Currently inlet tot flow is 51.3, split mode.  So at least the back flow is back to normal.  Will continue to work on this, but the back line was the priority.  We only use the front TCD once in a while, but any other ideas would be great.  Thanks

Reply
  • So the EPC board did fix the back flow, I had improperly zeroed the new board.  So now the back flow is hitting right on the nose.  The front flow is still low.  if I increase the oven temp from 50 to 100 the flow reading moves from 1.4 to 1.78 on the meter (which I did verify on another instrument).  Currently inlet tot flow is 51.3, split mode.  So at least the back flow is back to normal.  Will continue to work on this, but the back line was the priority.  We only use the front TCD once in a while, but any other ideas would be great.  Thanks

Children
No Data
Was this helpful?