In datasheet the gc 7890 writing about  posible instal 3 detector maximum, but in datasheet gc 8890 writing about  posible 4 detector. I never see 4 detectors are installed simultaneously on one chromatograph. for exmple in the gc7890 posible- MS///FID///FID install detector - total 3 detector. How/where to install 4 detectors on 8890GC ? I want to know. 

And if possible then the photo?

Fanks!

 

regards Azrael

Parents
  • Hi Azrael,

    8890A can install detectors up to 4.

    8890A has 8 slot for EPC. EPC map is below.
    Detector EPC can install slot 5-8.

     

    Slot Number       EPC Module
    1                         Back inlet; PCM configured as an inlet.
    2                         Front inlet; PCM configured as an inlet.
    3                         PCM A; PSD 1; AUX EPC 123; Helium Conservation module.
    4                         PCM B; PSD 2; AUX EPC 456; QQQ CC.
    5                         PCM C; PSD 3; AUX EPC 789; AUX Detector 2 (side).
    6                         PCM D; PSD 4; AUX Detector 1 (top).
    7                         Front Detector.
    8                         Back detector.