8890 GC - four detector

Hello !

I interesting gc8890 and gc7890. 

In datasheet the gc 7890 writing about  posible instal 3 detector maximum, but in datasheet gc 8890 writing about  posible 4 detector. I never see 4 detectors are installed simultaneously on one chromatograph. for exmple in the gc7890 posible- MS///FID///FID install detector - total 3 detector. How/where to install 4 detectors on 8890GC ? I want to know. 

And if possible then the photo?

Fanks!

 

regards Azrael

Parents
  • Hi Azrael,
    8890A has 4 position for detectors. I not sure side mount box has 2 detectors.

    1, Front position: all detector can install

    2: Back position: all detector can install

    3: Middle position(Oven top) : it can install only TCD and FPD+. See attached pic

                                                     Not compatible Valve Box and Dual FPD

     

    4: Side mount: it can install only FID, TCD, ECD and NPD. See attached pic, "Side mount.jpg"

                            Not compatible MSD

     

    Regards,

    Akihito Kumazawa

Reply
  • Hi Azrael,
    8890A has 4 position for detectors. I not sure side mount box has 2 detectors.

    1, Front position: all detector can install

    2: Back position: all detector can install

    3: Middle position(Oven top) : it can install only TCD and FPD+. See attached pic

                                                     Not compatible Valve Box and Dual FPD

     

    4: Side mount: it can install only FID, TCD, ECD and NPD. See attached pic, "Side mount.jpg"

                            Not compatible MSD

     

    Regards,

    Akihito Kumazawa