Volume Accuracy Failure - 850-DS Dissolution Sampling

During volume accuracy test of our 850-DS Dissolution Sampling, the difference of the sample volume for position #2 for 10mL was higher than 0.5mL, being that specification is lower than 0.25mL.

How can we calibrate the equipment to sample the right volume?

We performed the full range volume dispense (in the calibration screen), but we are not sure how to use the data to do the equipment calibration.

Could you help us, please?

# Empty Test Tube Filled Test Tube Difference Actual Sample Volume Difference from Set Point of 10 mL
1 7.6591 17.6115 9.9524 10.0226 -0.0226
2 7.6452 17.0647 9.4195 9.4859 0.5141
3 7.5446 17.4167 9.8721 9.9417 0.0583
4 7.6616 17.4081 9.7465 9.8152 0.1848
5 7.6411 17.5016 9.8605 9.9300 0.0700
6 7.6298 17.4850 9.8552 9.9247 0.0753
7 7.5995 17.4570 9.8575 9.9270 0.0730
8 7.5819 17.4926 9.9107 9.9806 0.0194

Thank you so much.

Suzany

Parents Reply
  • Hi Lycrecia,

    We checked the volume calibration section and we did the full range volume dispense, as described in the manual. However, we are not sure which value we should add to the actual volume in the volume calibration screen. Do we need to do the average volume for each range tested (0.1, 0.5, 2, 4, 10 and 15mL)?

    How can we do this for only one position (position #2) that failed? 

    Thank you!

Children
Was this helpful?