Current = 4.0 uA, Expected range: 5 - 1500
Current = 4.0 uA, Expected range: 5 - 1500
Thank you for contacting the Agilent team.
To address the error related to current, please address the points outlined below:
If issue persists, please email your local support with screenshot of the error, event and log reports. To collect event reports, follow the steps below:
To collect the Error Report, follow the same steps as above, but instead go to Admin > Error Report. The rest of the steps are the same.
Thank you, Kadir Tung
We tried all the above steps, but the problem continued. We had to change the parameters in the Methods file as suggested in another reply from Cendy D. to make the instrument continue
How do we avoid the file corruption, which is what we found in our files and needed changing? Our values were also 0 as shown in the Fragment analyser (HS NGS kit) fail to achieve 4µA ?
Hello,
This bug is known as the method save bug. This occurs when you are editing the method and do not follow the attached instructions below. Please note, that even if no change is made, but the method is opened in the Edit Method window and saved, the method will become corrupted. If you are purposely making changes to your method, please follow the attached guidance below.
If you do not intend to make changes to the methods, please consider logging in the software with the user level privilege, Username: User.
The default methods can be downloaded from our software website and copied to your methods folder to replace the corrupted method files.
Software Download Fragment Analyzer Systems | Agilent
Best Regards,
Ian