Low Li-7 tune higher on In-115 and U-238 on Agilent 7700

We have an agilent 7700. When we tune our Li-7 is coming in very low (2,300). Our In-115 and our U-238 are higher than before. What levels should we be getting for a 1ppb tune solution?  What is the best method to tune for improving our Li-7 signal? Can we tune just the extraction cell in manual tune?

Parents
  • Hi, 

    without any comparison it is hard to say what is going on. Have you changed anything and then this behavior occurred?

     If you want to focus the intrument performance for lower masses, you can set this by right clicking in the tune window and select "Tune - Set Acq Parameters for Sensitivity"

    There you have the option to set the mass balance:

    Do you only measure light elemtens? And why do you tune with In and U?

    Best Regards

    Chris

Reply
  • Hi, 

    without any comparison it is hard to say what is going on. Have you changed anything and then this behavior occurred?

     If you want to focus the intrument performance for lower masses, you can set this by right clicking in the tune window and select "Tune - Set Acq Parameters for Sensitivity"

    There you have the option to set the mass balance:

    Do you only measure light elemtens? And why do you tune with In and U?

    Best Regards

    Chris

Children
Was this helpful?