Dear All,
We would like to start using of bracketing technique in our ICP OES 5800. Do you know if there is any option here to choose or create this kind of method?
Dear All,
We would like to start using of bracketing technique in our ICP OES 5800. Do you know if there is any option here to choose or create this kind of method?
The ICP Expert software does not have a bracketing option on the standards page, however, you could try unchecking "Include blank in calibration" and then enter your standards as Calibration Standard 1 and Calibration Standard 2.
The ICP Expert software does not have a bracketing option on the standards page, however, you could try unchecking "Include blank in calibration" and then enter your standards as Calibration Standard 1 and Calibration Standard 2.