how to add a tune mode into a batch? I am using ICP-MS 8900
how to add a tune mode into a batch? I am using ICP-MS 8900
Hi Triumf,
If you use MassHunter just do the following: Make a batch file. Then go to the File menu and select Import from existing batch. Choose a batch that you used before and had an appropriate tune mode in it.
Unselect acq method/ DA method/ sample list and QC config. But instead select Tune mode. On the right you can then select one or more extra tunes to be copied in your new batch.
Good Luck!
Hi Triumf,
If you use MassHunter just do the following: Make a batch file. Then go to the File menu and select Import from existing batch. Choose a batch that you used before and had an appropriate tune mode in it.
Unselect acq method/ DA method/ sample list and QC config. But instead select Tune mode. On the right you can then select one or more extra tunes to be copied in your new batch.
Good Luck!