We get an interference next to mass 7 peak.

Almost every time we run the performance check we get this peak at mass 6, next to mass 7. We tried cleaning and/or changing skimmer plus cone, but we still get this interference. We use high quality nitric acid and milli-Q water. A while ago we had the omega lens cleaned as well, and it got a little better, but it only lasted two days! After two runs there it was again. Has this happened to anyone? We are grateful for any help or advice. Thank you! 

Parents Reply
  • Glad I could help.  This is a common issue.  The internal standard line is much smaller diameter than the sample line, so it takes a lot longer to flush it out.  The 15 minute warmup should be enough with a 1% nitric acid solution to clear it.  Even if there is a small peak there, as long as it is smaller than the Li7 peak it will find the correct peak.  Best Regards.